How to Calibrate hot air on Aoyue 968

Rework and Rework Station guides and tutorials.

How to Calibrate hot air on Aoyue 968

Postby jamesfreddie » Thu Nov 15, 2012 3:43 pm

The system is not calibrated such that the display would match the actual temperature sensor data as doing so would make the hot air gun unusable. This is because the sensor is located inside the hot air gun. As air is passed thru the heater and the sensor out thru the nozzle there would be temperature variations from the sensor and at the measured site. Normally the system is calibrated to have a display of 480c but an internal sensor temperature of 650-700 degrees. (This is to account for the heat loss due to distance from tip of nozzle to target component and the different thermal mass of various nozzles).





IMPORTANT: Follow all proper electrical safely measures when calibrating the system in live mode with the back panel opened for access to the calibration port. Only qualified personnel are allowed to open the back panel of the system to perform calibration procedures.



Caution: High voltages are present inside the system.



Simple Calibration procedures: (Factory calibration procedures requires an attachment to the internal sensor for precise calibration)

1. Unplug system from mains.

2. Loosen the screw which hold the back plate.

3. Locate the calibration port. Use a small but long philips screw to reach the calibration port, make sure not to touch or allow the screw driver to come in contact with any other components in the system.

4. Attach a 1124 or 1130 nozzle on the hot air gun. This setup would concentrate the hot air stream for better temperature readings.

5. Place a k type thermocouple 2-5 mm from the nozzle directly in the path of the hot air stream.

6. Plug system to power source.

7. Turn on the hot air gun function.

8. Adjust set temperature to 480 degrees, wait for temperature to stabilize.

9. Check the external temperature sensor reading, adjust the calibration port "W1" encircled in red in the attachment (968 hot air calibration.JPG).

10. Adjust calibration port such that the displayed temperature and the external temperature sensor reading is almost equal or within 10 degrees of each other.



Note: This calibration procedure disregards the thermal effects of other type of nozzle. If a larger nozzle is used than the actual temperature coming out of the larger nozzle would be lower than the displayed temperature. Different distances of the target component and different nozzles would affect this calibration. It is greatly recommended that the user not recalibrate the device but simply increase/decrease the set temperature settings to compensate for the different target component distances and nozzle sizes.

Image

Here is a Video to demonstrate :D

[youtube]http://www.youtube.com/watch?v=a_XXvtB93tg[/youtube]
jamesfreddie
Banned
Banned
 
Posts: 914
Joined: Thu Nov 08, 2012 6:28 pm
Location: Waltham Abbey, Essex, England
National Flag:
Great Britain
Has thanked: 58 times
Been thanked: 115 times
Set up: Honton R392 Atten 858D, Aoyue 968, T8280, SCOTLE HT-80, WEP 942 Aoyue 2901 Clone

How to Calibrate hot air on Aoyue 968

Sponsor

Sponsor
 

Re: How to Calibrate hot air on Aoyue 968

Postby Flashing360 » Wed Apr 10, 2013 12:39 am

topic split james and created your own guide on this ;)

good work mate.

:+1:
Flashing360
 

Re: How to Calibrate hot air on Aoyue 968

Postby DrunkenNinja » Wed Apr 10, 2013 4:28 am

nice one James I might just give this a go while I wait to mod my wand holder.
User avatar
DrunkenNinja
VIP Member
VIP Member
 
Posts: 1636
Joined: Mon Nov 05, 2012 9:30 pm
Location: Nicosia, Cyprus
National Flag:
Great Britain
Has thanked: 594 times
Been thanked: 475 times
Set up: Aoyue 968 - Modified Heatspreader / Puhui T-8280 - C100 Controller Modification

Re: How to Calibrate hot air on Aoyue 968

Postby evestu » Sun Jan 19, 2014 11:37 am

Thank you mines was out by 30oc now it is alot better :)
evestu
New Member
New Member
Progress to next rank:
80%
 
Posts: 4
Joined: Wed Jan 15, 2014 7:10 pm
Location: essex
National Flag:
Great Britain
Has thanked: 2 times
Been thanked: 0 time

Re: How to Calibrate hot air on Aoyue 968

Postby frankfiller » Wed Mar 11, 2015 1:26 am

Just did this, mine was of by 45°C. Thanks for the post.
frankfiller
New Member
New Member
Progress to next rank:
20%
 
Posts: 6
Joined: Tue Nov 18, 2014 11:14 pm
Location: United States
Has thanked: 0 time
Been thanked: 0 time

Re: How to Calibrate hot air on Aoyue 968

Postby drwcifer » Wed Mar 11, 2015 11:59 pm

Great idea..What do you put the air speed at?
drwcifer
VIP Member
VIP Member
 
Posts: 124
Joined: Sat Feb 01, 2014 11:25 pm
Location: Kansas
National Flag:
United States of America
Has thanked: 14 times
Been thanked: 2 times
Set up: Puhui T-8280 AOYUE 968A+ Rework Station

Re: How to Calibrate hot air on Aoyue 968

Postby sklickrick » Fri May 08, 2015 10:28 pm

[quote="jamesfreddie"]Attach a 1124 or 1130 nozzle on the hot air gun. This setup would concentrate the hot air stream for better temperature readings.[/quote]

maybe calibrating with the most common Nozzle used , would be better? dunno..

[quote="drwcifer"]Great idea..What do you put the air speed at?[/quote]

id also calibrate with the most common / average air speed! :D :D
sklickrick
Dedicated Member
Dedicated Member
Progress to next rank:
80%
 
Posts: 24
Joined: Tue Jan 20, 2015 2:48 am
Location: MANCHESTER
Has thanked: 0 time
Been thanked: 1 time


Return to Rework/Station Guides